I'm starting an ion exchange HPLC (previously done mostly RP-HPLC) so I could use some advice.
1. how to rinse system from the buffer in mobile phase - when the system is not use it is necessary to rinse it from the buffer so the pumps won't get damaged.. Is there a good way to do without rinsing the column? I'm thinking to have a flow switch valve so the system can get good rinse but not the column (so I do not have to wait for the conditions to equilibrate and also ion-exchange resins have the tendency to shrink/swell in different composition of the mobile phase so I'm worried about shrinking and expansion of the resin).
2. I'm using purely aqueous buffer, what would be a good rinsing mobile phase? According to what I found pure water is not recommended because of the possible microbial growth.. So a small amount of organic solvent is recommended like methanol.. How much methanol (other) solvent to use to prevent the microbial growth?
3. any other tips for newbie in IC-HPLC? Good literature to read...

Thanks!
Veronika
Hi Veronica,

Depending on how long the system is going to dwell (stay unused) I don't think you need to worry about bacterial growth if you use pure water and give the system a day or two off. After all an HPLC system is not the best thinable medium for a micro-organism to thrive in ;-)

You don't have to worry about stationary phase support shrinking/swelling either. It is usually seen when organic modifiers (like methanol etc.) are used - especially shifting from aqueous to organic and back.

But flushing the system with pure water is always a good idea after using salts - wich is the scope of your plans.

Best Regards
Learn Innovate and Share

Dancho Dikov
Dear Veronica

danko is right... anyway I would like to add a few points.

1. check your column leaflet about any change in solvent. Some columns do not like pure water at all.
2. Micro-organisms are able to grow anywhere! Therefore when stopping the system for longer time I recommend to finally keep it in e.g. 25% methanol.
3. Flushing the system with water: make sure that at any time the components are miscible.

Best regards
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Thanks for the suggestions!
V
4 posts Page 1 of 1

Who is online

In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Google Adsense [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry