Siltite Ferrules for GCMS interface - problems
Posted: Tue Apr 09, 2013 12:27 pm
Hi all,
My first post here - although I use this board frequently to read threads.
I am using siltite nuts and ferrrules for the GCMS interface as I am using high oven temperatures. Using an Agilent 5975C in EI mode.
I am new to using these siltite ferrules but have just installed a new Inferno column (30 m x 0.32 mm i.d x 0.25 um film) using 0.32 mm ferrules and silitite nut.
As I pump the ms down - everything seems fine. Turn on the transfer line (280 C) and the over (40C) - still fine.
Turn the oven up to 100C - snap! The column breaks at the point where is joins with the MS interface nut. This means the MS gets very cross indeed as the vacuum is lost and shuts down.
I tried it again and the same happened.
Gone back to using a vespel/graphite ferrule and the standard Agilent nut. It is pumping down now so we will see.
Anyone had anything similar with these nuts and ferrules?
Thanks for your time,
Louisa
Edit: Forgot to say I am following the instructions for the ferrules. Tighten until the ferrrule bites the coloumn, then tighten a tiny bit more. I get an excellent air tight seal - the vaccum is 1 x 10-5 with less than 4% air in the tune.
Louisa