In our lab we run all QUECHERS samples on our Agilent GCQQQ systems. We do not do any solvent exchange prior to analysis on GC. We use the MMI (multimode inlet) on the 7890 in a solvent vent configuration so as to vent off the ACN. This produces excellent results on the GCQQQ in MRM mode and allows us to hit LODs for most pesticides of 1ppb in the best matrices and 5ppb in the worst. The solvent vent mode does an excellent job of removing the ACN without much discrimination. It also does a decent job of keeping the thermally liable compounds relatively happy. Not sure what kind of inlets you have available but solvent vent is worth considering if you want to avoid solvent exchange.
We also have a few 6890s with specific detectors that we run in a pulsed spliteless mode and for the very limited screens that we do with them it works surprisingly well. Not sure how well this method would work for a large screen but for our purposes it works quite well.
Our columns on both systems last for over a year normally.
Thank you for your response
We also want to go for LVI and solvent venting.
We have an OPTIC-4 (ATAS GL) on a shimadzu QP2010 system.
Any advice on the Solvent venting, time line, temparatur, heat up rate etc?