Hi Peter,
thanks for your reply.
In your job, are you so clever and lucky that your integration parameters fit for 100% of all your samples analyzed on different HPLC with different HPLC columns?
Fortunately I no longer work in a laboratory where I have to follow SOPs. When I was working in such an environment the SOPs were written in such a way that they could be followed as written - i.e. without the operators having to adjust conditions and integration parameters as they went along. Usually this meant that for each analyte / matrix combination the sample prep, running conditions, column etc would be specified, and the integration parameters would be considered as part of the analytical method. All of these would be validated before any results could be issued.
If an analysis had to be run with different columns on different instruments its robustness to these changes would be validated, and any changes that were necessary for a particular instrument would be written into the method section of the SOP.
Our aim was always to have good enough chromatography for the automatic integration to be easily and repeatably able to find peaks and integrate them. A continual need to be adjusting parameters or integrating by hand was seen as a symptom of a method that needed to be improved.
If you are having to change integration parameters to get "accurate" integrations then how do your methods pass validation ? And how do you know that the integrations that you get after your manual adjustments are "accurate" - do you run a set of control samples and standards to check each time. Or do you use parameters that put the product composition where you expect it to be ?
Peter