Column for impurities in butadiene?
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
Does anyone know of a column that will seperate all of the impurities in butadiene? I am using an aluminum oxide but methyl acetyle and pentane sometimes coelute and sometimes are swamped out by 1,3-butadiene which is the main peak.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
I would suggest that a BMEA SCOT capillary column of 15m or 30m combined with a methyl silicone would solve your problem.
To separate monoolefin isomers from the diene isomers requires a very polar column. In using a polar column certain non-polar hydrocarbons may coelute (pentanes and hexanes, especially). These are not usually present in 1,3-BD.
A multidimensional application is usually dictated for 1,3-BD analysis.
You may call or email me if you wish to discuss this application. It can be done in less than 8 minutes.
Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823
814-359-5737 voice
814-359-5459 fax
rgeorge@sial.com
To separate monoolefin isomers from the diene isomers requires a very polar column. In using a polar column certain non-polar hydrocarbons may coelute (pentanes and hexanes, especially). These are not usually present in 1,3-BD.
A multidimensional application is usually dictated for 1,3-BD analysis.
You may call or email me if you wish to discuss this application. It can be done in less than 8 minutes.
Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823
814-359-5737 voice
814-359-5459 fax
rgeorge@sial.com
-
- Posts: 36
- Joined: Sun Jan 16, 2005 12:36 am
Hi
you can use Alumina oxide/Na sulfate or manganate capillary ( 0.32 mm) or wide bore (0.53 mm) column.and helium as carrier.
with good temperature programing and constant flow you can analyse down to 1 ppm of impurities in 1,3 Butadiene.
we do this.
you can use Alumina oxide/Na sulfate or manganate capillary ( 0.32 mm) or wide bore (0.53 mm) column.and helium as carrier.
with good temperature programing and constant flow you can analyse down to 1 ppm of impurities in 1,3 Butadiene.
we do this.
3 posts
Page 1 of 1
Who is online
In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 13 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 13 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science