pressure drop during the run

Posted: Thu Apr 04, 2013 1:41 am
by xiuminchen
Hi,
I have a problem of our Agilent 1200 HPLC. The pressure is stable when I don't inject the sample, but after injection (around 5-6min) the pressure suddenly drops to almost 0 and increases again soon. Does any one know why is that?
thanks
xiumin

Re: pressure drop during the run

Posted: Thu Apr 04, 2013 1:50 am
by KM-USA
Post your gradient conditions.

If isocratic, you've really got issues !!!

Re: pressure drop during the run

Posted: Thu Apr 04, 2013 2:59 am
by xiuminchen
It is isocratic. And there is a leak from the bottom of active inlet valve. I changed pump seal and inlet valve cartridge.It still leaks.

Re: pressure drop during the run

Posted: Thu Apr 04, 2013 3:13 am
by xiuminchen
Should I change the Active Inlet Valve too?

Re: pressure drop during the run

Posted: Thu Apr 04, 2013 3:24 pm
by Consumer Products Guy
It is isocratic. And there is a leak from the bottom of active inlet valve. I changed pump seal and inlet valve cartridge.It still leaks.
Yes, active inlet valves can leak over time; you state "leak from the bottom of active inlet valve".

Also, sometimes the two pistons in the pump head get scratched and need replacing. I'll assume that you actually changed both piston seals. Also, the outlet valve can leak and trickle down. And did you use a new gold seal when doing that active inlet valve cartridge?

If your leak is truly "from the bottom of active inlet valve", replace that. It "ain't gonna" heal iself.

Re: pressure drop during the run

Posted: Thu Apr 04, 2013 5:18 pm
by xiuminchen
thank you very much. I will change the active inlet valve. When you say gold seal, do you mean the gold seal in the purge valve or in the outlet ball valve? I change the purge valve gold seal but haven't change the outlet ball valve gold seal.

Re: pressure drop during the run

Posted: Thu Apr 04, 2013 5:50 pm
by xiuminchen
Can anyone tell me why the pressure drop only happens when I inject the sample? I just did the leak test, every thing is fine.

Re: pressure drop during the run

Posted: Fri Apr 05, 2013 6:03 pm
by uzman
Check and replace rotor seal.

Re: pressure drop during the run

Posted: Wed Apr 10, 2013 1:01 am
by xiuminchen
we just changed rotor seal and I ran the sample on the following day, the leak came out from active inlet valve again. So, I will change active inlet valve and see whether it will solve the problem.

Re: pressure drop during the run

Posted: Thu Apr 18, 2013 5:33 pm
by xiuminchen
Hi, the problem solved after I change active inlet valve.