pressure drop during the run
Posted: Thu Apr 04, 2013 1:41 am
by xiuminchen
I have a problem of our Agilent 1200 HPLC. The pressure is stable when I don't inject the sample, but after injection (around 5-6min) the pressure suddenly drops to almost 0 and increases again soon. Does any one know why is that?
thanks
xiumin
Re: pressure drop during the run
Posted: Thu Apr 04, 2013 1:50 am
by KM-USA
Post your gradient conditions.
If isocratic, you've really got issues !!!
Posted: Thu Apr 04, 2013 2:59 am
by xiuminchen
It is isocratic. And there is a leak from the bottom of active inlet valve. I changed pump seal and inlet valve cartridge.It still leaks.
Posted: Thu Apr 04, 2013 3:13 am
by xiuminchen
Should I change the Active Inlet Valve too?
Posted: Thu Apr 04, 2013 3:24 pm
by Consumer Products Guy
Yes, active inlet valves can leak over time; you state "leak from the bottom of active inlet valve".
Also, sometimes the two pistons in the pump head get scratched and need replacing. I'll assume that you actually changed both piston seals. Also, the outlet valve can leak and trickle down. And did you use a new gold seal when doing that active inlet valve cartridge?
If your leak is truly "from the bottom of active inlet valve", replace that. It "ain't gonna" heal iself.
Posted: Thu Apr 04, 2013 5:18 pm
by xiuminchen
thank you very much. I will change the active inlet valve. When you say gold seal, do you mean the gold seal in the purge valve or in the outlet ball valve? I change the purge valve gold seal but haven't change the outlet ball valve gold seal.
Posted: Thu Apr 04, 2013 5:50 pm
by xiuminchen
Can anyone tell me why the pressure drop only happens when I inject the sample? I just did the leak test, every thing is fine.
Posted: Fri Apr 05, 2013 6:03 pm
by uzman
Check and replace rotor seal.
Posted: Wed Apr 10, 2013 1:01 am
by xiuminchen
we just changed rotor seal and I ran the sample on the following day, the leak came out from active inlet valve again. So, I will change active inlet valve and see whether it will solve the problem.
Posted: Thu Apr 18, 2013 5:33 pm
by xiuminchen
Hi, the problem solved after I change active inlet valve.