Method parameters - the various settings for the GC method. Inlet and Oven conditions. I assume that you did not change carrier gas while shortening the column? And I assume column parameters are part of the ChemStation method? (It's been a while since I used ChemStation - perhaps on instruments too old to have electronic flow controllers.)
Here is the method with the
60 m column:
ALS
Injection volume: 1uL
Washes and Pumps: Sample Washes 3; Sample Pumps 6; Pre injection solvent (hexane) washes 2; post injection solvent (hexane) washes 2
S/Sl Inlet
Mode: split, split ratio 2:1
temperature: 280c
Pressure 17 psi
septum purge flow: 3 mL/min
Oven Temperature program: 50c, hold 1 min; 25c/min to 175c; 4c/min to 230 c, hold 10 min
Carrier gas, carrier gas pressure: He, 17 psi, constant pressure mode
MSD
Source T: 230c
Quad T: 150c
solvent Delay: 6 min
EMV Mode: Gain Factor
Acquisition Mode: Scan
Scan Speed: Normal
Low mass: 50.00
High mass: 550.00
Thermal Aux 2 T: 240c
Run Time: 29.75 min
Now a few changes for the
30 m column:
ALS
Injection volume: 1 uL
Washes and Pumps:
sample washes 6 (not sure how or why that changed); sample pumps 6;
pre injection solvent (hexane) washes 6 (again not sure how or why that changed); post injection solvent (hexane) washes 2
S/Sl Inlet
Mode: Split, split ratio 2:1, 2 mL/min
Temperature: 250c
Pressure: 7.6165 psi
Septum purge flow: 3 mL/min
Oven Temperature Program: 50c, hold 1 min; 25c/min to 175c; 4c/min to 230c,
hold 5 min (hold for 5 min is the only thing I changed about the T programming/ramping - I need to do more research on a T program method)
Carrier gas, carrier gas pressure: He, But instead of constant pressure mode,
constant flow mode was chosen at 1 mL/min; average velocity listed 36.445 cm/sec.
MSD
source T at 230c
quad T at 150c
solvent delay now at 4 min
Thermal Aux 2 Temp: 230 c
Run Time 24.75 min
Changes are in
BOLD