Page 1 of 1
TOGA help needed!
Posted: Tue Apr 02, 2013 12:38 pm
by akar
Hi! I am looking for transformer oil dissolved gas analyzer (TOGA) solution. I have proposal with capillary columns (HP-PLOT 5A 30m, 0.53mm, 50.0um + GS-CarbonPLOT 30m, 0.53mm, 3.00um) and another proposal with packed columns. Which solution is better, packed or capillary for TOGA?
Re: TOGA help needed!
Posted: Tue Apr 02, 2013 1:59 pm
by CE Instruments
I would always go capillary as separation should be better but either column system should be OK. The real differences are in the extraction methods, my favourite is with Headspace and D3612 method C.
If you are in the UK please get in contact via our website (link in profile)
Posted: Wed Apr 17, 2013 1:47 pm
by jdezeeuw
Both GC configurations will work.
your challenge is the sample preparation.
I have seen several "capillary" solutions, but in-time they contaminate becuase of the residual oil. You need a could pre-separator for that.
If this is not assured, go for the packed solutions as that can take more "stress" before giving up.
jaap de zeeuw, Restek corporation
Posted: Mon Apr 22, 2013 4:15 am
by thohry
I have run transformer oil analysis by GC-Headspace with capillary column. But I see some labs use stripper column for extracting the dissolved gases.
Anyone who have run on stripper column configuration could say which one is more advantageous ?
Posted: Wed Dec 11, 2013 10:01 am
by oiltest
Producer and exporter of Transformer Oil Dissolved Gas Analyzer in China.
Conforms to ASTM D3612 and IEC 60567.
High quality and Competitive price.
For more information, please contact us through
www.dgaequipment.com.