I am running a pulsed splitless injection. I just recently started using a liner with glass wool. The first 10 analytes have peaks that are splitting or have jagged tops. What can I do to eliminate this?
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
Try less glass wool. If that reduces the splitting, try increasing the pulse time.
Is your inject temperature well above the solvent boilng point or is it a cold injection?
Is your inject temperature well above the solvent boilng point or is it a cold injection?
-
- Posts: 2
- Joined: Wed Apr 03, 2013 4:09 pm
The injection port is at 280 at injection.
I just tried increasing the pulse time from 0.25 to 0.30 and there was no change.
I am using liners that come with the wool already packed. I'm not sure that if i took some wool out that i could be consistent.
I just tried increasing the pulse time from 0.25 to 0.30 and there was no change.
I am using liners that come with the wool already packed. I'm not sure that if i took some wool out that i could be consistent.
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
Do you have a retention gap (short piece of uncoated column) or other extra volume?
My first thought was that you are getting some extra retention in your inlet, a leak or extra volume would do that. If you have solvent condensing at the head of your column it can cause problems with peak shape in early eluters.
If you try it with a small split does the problem go away?
Does your liner have enough volume to hold the vapor on injection? A different liner with wool may not behave the same.
My first thought was that you are getting some extra retention in your inlet, a leak or extra volume would do that. If you have solvent condensing at the head of your column it can cause problems with peak shape in early eluters.
If you try it with a small split does the problem go away?
Does your liner have enough volume to hold the vapor on injection? A different liner with wool may not behave the same.
4 posts
Page 1 of 1
Who is online
In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 29 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 29 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science