RP ghost peaks
Posted: Sat Mar 30, 2013 10:09 pm
I am trying to resolve a persistent set of ghost peaks in a simple reverse phase method. I am using a Water 2695 with dual wavelength detector at 215nm. Mobile phase A is UHQ, 0.1% TFA, mobile phase B is Acetonitrile, 0.1% TFA. The equilibration phase is 80% mobile phase A, and immediately after injection it starts and up to 50 minutes it ramps up to 100% mobile phase B, then drops back for 10 minutes to 80% A, then back to injection.
The ghost peaks although small (0.004-5 AU) are persistent and unfortunately coelute with a minor component of a drug product I am analysing. I know if I equlibrate for longer durations than specified I can elevate the levels of these ghost peaks on the first injection, on subsequent injection that are consistently present but at a much lower level. When I condition the column (ie: no injection but run gradient) I still get these peaks so my assumption is a minor contaminant in the mobile phase collecting on the column during equlibration then washing off at a sweet spot of organic during the gradient, then building up margainally on subsequent cycles.
I have however, sourced 4 different water sources, 5 different column batches, 4 different acetonitrile sources, attempting different glassware cleaning regimes, purchased in pre-made mobile phases, etc and I cannot shift the appearence of these peaks. HELP !
The ghost peaks although small (0.004-5 AU) are persistent and unfortunately coelute with a minor component of a drug product I am analysing. I know if I equlibrate for longer durations than specified I can elevate the levels of these ghost peaks on the first injection, on subsequent injection that are consistently present but at a much lower level. When I condition the column (ie: no injection but run gradient) I still get these peaks so my assumption is a minor contaminant in the mobile phase collecting on the column during equlibration then washing off at a sweet spot of organic during the gradient, then building up margainally on subsequent cycles.
I have however, sourced 4 different water sources, 5 different column batches, 4 different acetonitrile sources, attempting different glassware cleaning regimes, purchased in pre-made mobile phases, etc and I cannot shift the appearence of these peaks. HELP !