Perplexing compound
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
I have been asked if I can make an LC/MS/MS method for bromoxynil, it's been around for ages but usually analyzed by other means. The problem is the target is the heptanoate and bromoxynil heptanoate (around 389 da) isn't what you see in the first quad it's bromoxynil (276 da). Why do I see the 276-80 transition elute with multiple peaks late in the gradient? I know due to stability and the ester that if you don't acidify samples it will all be bromxynil alone and elute early due to it's polarity but I don't want that. I can acidify the sample and stabilize the compound, remove the early eluting peak but can't find a decent peak that I can use to quantitate with. I want the non-polar late eluter so it's undeniably bromoxynil heptanoate and not just bromoxynil. I see some late eluting peaks but they drift and I'm not sure what I'm looking at. Nasty adsorption too, I've run into very few compounds that I've actually had carry over with and this is one of them. I'm thinking maybe pass this compound off to the GC or HPLC guys and let them extract it if they have to. Does anyone have a trick or done this and can say yeah, use this column and mobile phase and it will work beautifully? I'm currently using a 50 mm C18 and CH3CN H2O formic mobile phases.
-
- Posts: 357
- Joined: Wed Aug 03, 2005 1:15 pm
(1) BROMOXYNIL and (2) BROMOXYNIL OCTANOATE 5010
www.cdc.gov/niosh/docs/2003-154/pdfs/5010.pdf
gives a very straightforward HPLC method (C18/ MeCN/ Water/ 254 nm) with RT 3 min and RT 33 for the two analytes, respectively.
The heptanoate would be expected to elute ahead of the octanoate.
Note the cautions "refrigerate and protect from light" for the aq. samples that are to be analyzed.
www.cdc.gov/niosh/docs/2003-154/pdfs/5010.pdf
gives a very straightforward HPLC method (C18/ MeCN/ Water/ 254 nm) with RT 3 min and RT 33 for the two analytes, respectively.
The heptanoate would be expected to elute ahead of the octanoate.
Note the cautions "refrigerate and protect from light" for the aq. samples that are to be analyzed.
-
- Posts: 46
- Joined: Fri Jul 20, 2012 9:24 pm
Thanks, I know just what's going on now.(1) BROMOXYNIL and (2) BROMOXYNIL OCTANOATE 5010
http://www.cdc.gov/niosh/docs/2003-154/pdfs/5010.pdf
gives a very straightforward HPLC method (C18/ MeCN/ Water/ 254 nm) with RT 3 min and RT 33 for the two analytes, respectively.
The heptanoate would be expected to elute ahead of the octanoate.
Note the cautions "refrigerate and protect from light" for the aq. samples that are to be analyzed.
3 posts
Page 1 of 1
Who is online
In total there are 81 users online :: 0 registered, 0 hidden and 81 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 81 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 81 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science