Void Volume question

Posted: Tue Mar 26, 2013 6:57 am
by Leni
Please help me settle an argument.
Is the void volume/time of the column taken into account by HPLC software when calculating system suitability parameters(resolution, tailing, theoretical plates etc.) and if yes, then how/when?
Empower asks for void volume in a mandatory way to calculate system suitability parameters, but in chemstation these calculations are made even if you don't enter any void volume (or is a default value used?!)
I wasn't really sure, so i didn't want to argue if i didn't know for sure.
Please help,
Thanks...regards, Leni

Re: Void Volume question

Posted: Tue Mar 26, 2013 12:43 pm
by HPLCaddict
For calculation of the "usual" suspects regarding SST, i.e. tailing factor, plate count and resolution, the void time t0 is NOT needed. Have a look at the formulas, e.g. here:

http://www.separations.us.tosohbioscien ... ations.htm

t0 is needed for calulation of capacity factor (retention factor) k, which might sometimes be a parameter also included in SST.

Re: Void Volume question

Posted: Thu Mar 28, 2013 11:34 am
by Leni
For calculation of the "usual" suspects regarding SST, i.e. tailing factor, plate count and resolution, the void time t0 is NOT needed. Have a look at the formulas, e.g. here:

http://www.separations.us.tosohbioscien ... ations.htm

t0 is needed for calulation of capacity factor (retention factor) k, which might sometimes be a parameter also included in SST.

Thank you,
yes i have also conducted some research and came up with the same conclusions. Thanks again.