Related substance by HPLC
Posted: Sun Mar 24, 2013 8:01 pm
by mohammad
Dears,
As per the EP monograph for related substances by HPLC for any molecule.
Do we need to check the %RSD of refernce solution diluted from test solution (Example : 0.1% or any other concentration with respect to test concentration). If yes, how many injections. & what wil be the limit for RSD.
Since the EP says just inject refernce solution.
Any refernce shall be highly appreciated.
Mohammad.
Re: Related substance by HPLC
Posted: Thu Apr 25, 2013 11:54 am
by Perreman
EP usually states some sort of system suitability, you should be able to find the information on what to relate your reference solution to. It's often a %RSD criteria for 6 injections, a resolution criteria between two peaks or something like that.
What monograph are you using?