Dear all,
I applied the Gradient Performance Checks described by Gilroy and Dolan in LC.GC Europe 17(11) 2004 on my quaternary, low pressure mixing gradient pump. The reason for doing these checks was to identify the cause of getting lower resolution and retention times when mixing the eluent in the pump through channel A (water) and B (acetonitrile) compared to what I get normally by using a premixed eluent (40% Acetonitrile/Water).
In the Gradient Proportioning-Valve test (GPV) (Channels A, B= Water, C, D=1%Acetone), the average difference between the maximum and minimum plateau heights was 4.4% which is close to the 5% allowable maximum mentioned in the article.
For the Linear Gradient Test, I got two slopes for the gradient line when running Channel A=Water, B=Acetone. The line started steeper to about 20% B and less after that to the end of the line.
When doing the Step Test for channel A=water, B=Acetone, I got steady baseline at about zero for time lengths of about two steps then a high step but with no clear plateau which represents the 30% B (I think). The chromatogram is shown in this link:
http://www.albumtown.com/showpic.php?ai ... pid=516373
I thought that one of the two channels at least must have a problem. Therefore, I repeated the step test with the other channels as follows:
Water/ Acetone
A/ C
A/ D
B/ D
D/ B
However, all the other previous channel combinations produced clear steps from 10% to 100%. Despite that, all the combinations generated lower practical %B than the theoretical programmed %B. In addition, all the combinations also failed to achieve less than the maximum allowable1% step size difference (the practical step sizes were lower than the programmed from 0%B to about 70%B and higher afterwards).
I am wondering about the following:
What may have caused the previous effect in all the three tests?
Why only A vs B channels had the fused steps at the beginning?
What can I do to solve this problem? (Throw the pump?!
)
Thank you all in advance and sorry for this LONG STORY!