Im new to the forum so hello all!
My lab have purchased a UPLC system. Following my training course i have been assigned a method redevelopment on uplc.
My problem is this:
Started a method development for an analyte. I quickly developed an isocratic method which is fine however lacking the sensitivity for my target LOQ (0.2ppb!) i decided to explore gradient elution however my attempt at a scouting gradient has failed. The gradient was eluting large interferring peaks over the entire course of the gradient. The highest peak fluorescing at 300EU! So investigating this i have ran the gradient without an injectionand obsereved the output - same profile. Keeping in mind this is a brand new system and column! I therefore suspect one of the two mobile phases (ACN and 0.1% formic acid). I prepared fresh mobile phases, flushed the system and column in MeOh. Still no different. I subsituted the formic acid for TFA and the contamination lowered significantly and continued to decrease with each successive column condition. The interference is swallowing my peak.

Is this a normal occurance using fluoresence detection and gradient? All reagents are hplc grade and the two modifiers are above 99% pure. The water is from a milli-q system and im starting to think its the water that is contaminated? I have asked for some water to be ordered but it wasnt met with approval.

Hope someone can help!

Conor
Check out the "chalk talk" on gradient artifact peaks for a number of suggestions:
viewtopic.php?f=31&t=19085
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 2 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google Adsense [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry