- Posts: 50
- Joined: Tue Oct 30, 2012 2:26 pm
For example, the pressure varies between 42-46bars (with more or less regularity) for a system were it's usually between 45-46bars.
Any ideas?
Thanks
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Add to that the filter frits on the solvent lines. Especially problematic is any that sit in aqueous solutions. If algae grows on the frit it will have problems pulling solvent through and cause many problems that seem to come from nowhere(know this too well from personal experience). Also make sure no algae has managed to grow in the degasser, it can give you fits too, and very difficult to get rid of.Pressure fluctuations are typically caused on this unit by:
bad outlet valve
bad active inlet valve and/or cartridge
pump seals/pistons
improper purge
improper degassing
fluctuating/malfunctioning multi-channel gradient valve (even if running isocratic)
The last item can be bypassed, or all four leads put into one pre-mixed reservoir, lines purged, then check pump.
Pressure fluctuations are typically caused on this unit by:
bad outlet valve
bad active inlet valve and/or cartridge
pump seals/pistons
improper purge
improper degassing
fluctuating/malfunctioning multi-channel gradient valve (even if running isocratic)
The last item can be bypassed, or all four leads put into one pre-mixed reservoir, lines purged, then check pump.