Testing CF4 by FID GC

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
Hello,

I'm trying to test carbon tetrafluoride (R-14) by FID GC, but I can't get any sort of significant peak. The area of the curve is about 100 at best, and I've tried it on multiple FID instruments with the same result. I'm regulating the pressure down to about 20 psi and using a 6-port autosampler.

Any thought would be much appreciated.
I doubt that this can be done; an FID responds to C-H bonds and to a lesser extent C-C bonds. Your molecule has neither.

Peter
Peter Apps
I am not surprised by the lack of response. I suppose you are using capillary columns.

Usually packed columns with a much larger sample size are used if a FID is employed.

Do you have any other detector available?

Rod
I have a TCD/DID, but because CF4 elutes so close to nitrogen on the TCD, I wanted to confirm that is UHP CF4, since N2 won't show up on the FID. I'm using a ss packed column right now.
Are you using a MS13X? Or some other packing? Like a porous polymer?

Rod
5 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry