I am looking for a method for separating brassicasterol, campesterol, stigmasterol, and B-sitosterol. I did find a partial method that uses a C18 column, but the stigmasterol and the campesterol co-eluted (if i remember correctly). Any help would be a great help!

Doug

Did you already try a C8 column? Campesterol has one carbon less the sitosterol and stigmasterol has a double bond more then sitosterol. rom my experience these compounds will most likely elute together on a C18 column.

How are you detecting these sterols? Their UV response is not very good. Perhaps you do not require a sensitive method?

Sterols like this can be separated very easily by GC with much higher efficiency than HPLC; they work better if you make the trimethylsilyl derivatives. Maybe you don't want to go down this road though.

I plan on using a ELSD for detection. I have already separated them by GC, but I need a small amount of one of the sterols and was hoping to be able to separate it out of a mixture.
I have been fighting with sterol separations for quite a while...
My problem is related to phytosterol esters (fatty acid esters) in combination with fish oil included in the matrix.

I did have one try where I hydrolized my phytoesters before analysis which simplified my chromatogram, however it didn't help because these components co-eluted with fish oil components. To cut to the chase however, I got good separation of the free esters from each other with these conditions.

M.P.
A = 5%ACN/95%IPA
B = ACN
60%A:40%B Isocratic
Column oven @45C
UV detection @204nm
Phenomonex Gemini C18 5u 4.6X250

The K' of most of these peaks is proportional to the amount of ACN you have, you can increase-decrease resolution fairly easily by adjusting the %B component.

Also If I remember correctly there is an application in one of the back issues of LC-GC's application notebook for separating free sterols... Unfortunately I don't recall which issue.
5 posts Page 1 of 1

Who is online

In total there are 24 users online :: 3 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], kaladey84, Semrush [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry