You should cap each vial immediately after the benzene solution is added, not after you have added it to the six vials.
You should have your benzene solution in a non-polar solvent, instead of water, if that is the case.
You should have exellent RSD values, around 2% unless your technique is poor.
best wishes,
Rod
i'm still confused, i have operated just as you say, long time ago. even two people coorperate, one transfer liquid, the other cap.
i'm do as USP467 say, it's forbided not do as usp. by the way,the other components such as methanol,acetone,isopropanol,furan, i can reach less than 1% for rsd of area occationally. just the benzene hardly to pass the criterion of 15% rsd。
let me say the solution process:100ml volumetric, add 10ml DMSO,then weigh 40mg benzene into it, dilute with water to volume, and mix,400ppm.Transfer 1.0 mL of this solution to a 200-mL volumetric flask, previously filled with about 90 mL of water, dilute with water to volume, and mix,2ppm. Transfer 5 mL of this solution to a 100-mL volumetric flask, previously filled with about 50 mL of water, dilute with water to volume, and mix,o.1ppm.
i have started this validation from july,so far,not amazing development,drive me crazy. could u give me sone excellent suggestions to resolve the problem?thanks for ur attention.
best wishes,
dogcatlike