- Posts: 5
- Joined: Thu Jan 24, 2013 5:00 pm
Here's what I've looked at:
- Changed column from 30m DB-1 w/ 250um diameter, 0.25um film thickness
to 20m DB-1 w/ 180um diameter, 0.18um film thickness.
- Have good moisture, O2, N2 filtration
- Tunes to PFTBA are equivalent to or better than with Helium
- Cleaned source, repeller, and lenses
- Changed column and carrier gas type in ChemStation
- Right now it's set at 1 mL/min initial flow, 69 cm/sec average velocity, and 15 psi initial pressure, but I have tried running
at 0.5-0.7 mL/min flows that bumped the velocity down around 40-45 cm/sec. It lowered the baseline, but no improvement
in sensitivity.
We run some known mixtures monthly and after service to evaluate the performance of the instrument. On this particular instrument the abundances of the mixtures come out around 4-4.5 million. After the conversion, I get a broad bump for one of the peaks in the mixture and it's only 5k above the baseline, ~15-20k abundance total. I don't even see the other peaks in the TIC.
I'm not sure where to go from here. From some searching I've seen people suggest a different drawout plate in the source. Any recommendations or more info needed?
Thanks!
Dan