GPC detector problems

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi,
We have an GPC Waters Breeze System(Isocratic HPLC pump 1515; 717 plus Autosampler, 2414 Refractive Index detector and 3 columns Styragel HR1xHR 0.5xHR0.5; Flow rate = 1 ml/min).In this moment we changed 2 valves on the detector:Solenoid valve and Relief valve, because it flows in the same time on the purje mode and on the analysis.Four mounths after, this the GPC system it works, but after it begining again .What's happening? In this moment we change the flow rate= 0,5 ml/min, and it works normal(p=400 psi) .If I try to grow up the flow rate to 1 ml/ min,it flow again. Please help me.
Thank you,
Ileana
Check the detector outlet tubing ( waste tubing ) if it is clogged or not.
Also the inside diameter of this tubing is important for RI detectors.

Mostly , 0.5mm or greater ID tubing must be used , in order not to apply backpressure to the cell and solenoid valves , and also the tubing must not be too long.
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 3 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry