Maximum concentration for LC/MSMS !??????
Posted: Wed Mar 06, 2013 3:45 pm
Hi everyone, it's me again >^<
I am currently doing pesticide analysis via LC/MSMS, it's petty easy to do the MS tuning.
The main problem is that the calibration is a kind of not quite stable...
I already finish 7 pesticide standard, and 3 of them is not quite good...geez...
Those 3 pesticide are Fenitrothion, Quinalphos and Diazinion.
For Fenitrothion and Quinalphos, both the MS tuning is quite ok.
(one is around 10^4 respond, another is 10^5 respond)
The calibration is okay from 10 ppb to 50 ppb, however,
peak area of 100 ppb sample drops in those 2 standard,
peak area drop more obviously in higher concentration (150 ppb and 200 ppb).
I think it may due to decomposition at the first place but it seems not quite reasonable.
And for Diazinon, ah...I really get lost of what I should do,
the respond is very nice and give signal up to 10^7,
I predicted that the calibration can be easily done,
but sadly, the signal doesn't vary much with concentration.
Is it related to selective problem?
Should I just accept that calibration because the signal does increase with concentration?????
By the way, the system I used are Waters UPLC with Waters TQ detector.
Mobile phase is mainly around 85% ACN : water with buffer and acetic acid.
I will further optimize the flow rate after the individual calibration....
I need to thank you all first~! Because I know I must learn something here!!
I love this forum and am glad to be one here.