Hello--feel free to review this...don't know if it will work for MS/MS, but it's a nice paper. Don't need to use TFA for this...see what you think.
http://www.sciencedirect.com/science/ar ... 7399008638
Journal of Chromatography A, Volume 870, Issues 1–2, 18 February 2000, Pages 245–254
Ion-pair chromatography on a porous graphitic carbon stationary phase for the analysis of twenty underivatized protein amino acids P Chaimbault, K Petritis, C Elfakir, M Dreux
Institut de Chimie Organique et Analytique (IC.O.A), CNRS UPRES-A 6005, Université d’Orléans, BP 6759, 45067 Orléans Cedex 2, France
Abstract
The analysis of twenty underivatized protein amino acids has been achieved on porous graphitic carbon packing material (Hypercarb). Five perfluoroalkyl carboxylic acids (trifluoroacetic, heptafluorobutyric, nonafluoropentanoic, tridecafluoroheptanoic and pentadecafluorooctanoic acid) have been studied as ion-pairing reagent. Several parameters (equilibration time, quantities adsorbed onto the chromatographic support, concentration and nature of the ion-pairing reagent, as well as temperature effect) have been studied leading to the complete separation of these compounds in gradient elution mode. Evaporative light scattering detector has allowed the detection of these non UV–visible absorbent molecules. The chromatographic methodology developed can also be easily coupled with pneumatically assisted electrospray mass spectrometry
Especially See Fig. 5. LC–ELSD analysis of the 20 underivatized amino acids in gradient elution on Hypercarb (100×2.1 mm I.D.). Eluent A: nonafluoropentanoic acid (NFPA) 20 mM in an aqueous mobile phase. Eluent B: Acetonitrile. Gradient profile is: from 0 to 15% ACN in 10 min, then 26% ACN in 10 min and finally 50% ACN in a further 10 min. 50% ACN is maintained until the end. Column temperature is 10°C. Flow rate: 200 μl min−1. Detection: Evaporative Light Scattering Detector (ELSD) set as in section 2.2. Imp: Impurity
Matt