Adding an ELSD to Star Workstation control

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

7 posts Page 1 of 1
Hi All,

I am running Star Workstation 6.2 for a legacy HPLC system. In between the existing pumps and UV/vis detector is some sort of ADC board. I have acquired an Alltech 500 ELSD and am trying to figure out how to get the output into Star, ideally with the UV detector still online for dual detection. Though I could take the analog output of the ELSD and just swap it for the UV on the ADC board, does anyone know how to configure it so I can have both? I am a neophyte to the configuration of this instrumentation, but any help would be most appreciated.

Thanks, Nicole
Hallo

What is the Hardware of your HPLC system? Which pumps, detector .....are you using in this system?

Best regards
Chris
Hi Chris,

There are two HPXL pumps, one slaved. They run via ribbon cable into a non-descript box that I assume is some sort of ADC board, because it also has terminal lugs that accept as input the output from the Dynamax UV-1 detector, and for the injector contact. I am just hoping to somehow get the Star software to see the output from the Alltech detector (ideally) without disconnecting the UV-1. Thanks a bunch for taking a look at this :)
Nicole
Hello

You will need some D/A converter vor reading the signal of the ELSD in the software.
I think the box you did describe is a CIM-Box, it

right side on this picture:
http://www.labequip.com/rainin-rabbit-d ... ector.html

maybe you can connect the analog signal of your ELSD here.

Best regards Chris
Hi Chris,

I am not sure I fully follow you. The box I have does indeed look like the one in the pic you referred me to, but I always figured that was some sort of ADC, because the DC output of the UV-1 detector is already going into it. As you had suggested, In indeed plan to run the DC output of the ELSD into the same box, but there real issue is how to get STAR workstation software configured to even see this additional input.

Best,
Nicole
Hello

In system control, you can find a button "setup" in the prostar/dynamx window.
I think, in this window, you should be able to configure the CIM box. You will have to add a data acquistion/report block in the Method for the second channel to record the data.
But I am not sure if your CIM box has dual or single channel A/D, maybe the labels of the connectors can tell you more.

Best regards
Chris
Thanks Chris, we will give it a try
Nicole
7 posts Page 1 of 1

Who is online

In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry