Hi All.

Does anybody have a custom field for standard verification?
Example :
mean area Standard 1 (n=5)/mean area standard 2 (n=2) * wt std2/wtstd1 *100

Im new to Empower but aware you require labels in sample set to trend data

I've set up a suitable System suitability report tabulating precision, bracketing standards, USP plate, tailing etc, but custom field for above calculation is not as easy as I thought

Any help much appreciated

A
Long time since I used it but I think you set the first one as a standard (to calibrate against) then the second one as a "check standard" and it does the rest for you
Where can I buy the kit they use in CSI?
Thanks your reply.

Im not sure what you mean by check standards.
I thought you required a custom field to do this task.
Long time since I used it but I think you set the first one as a standard (to calibrate against) then the second one as a "check standard" and it does the rest for you
I think you mean "Control".
But this way you check your amount of your control against the known value, by using the first one as standard.

The OP wants to calibrate on both and then check the difference.

Ace
Thanks for reply,

Im still abit confused . Do I name std 1 (n=5) as control and std 2 (n=2) as standards?
how do I factor in the standard weights?
Thanks for reply,

Im still abit confused . Do I name std 1 (n=5) as control and std 2 (n=2) as standards?
how do I factor in the standard weights?
You use std (1) as standard, and std 2 as control.
Then you can use the amount field to factor in the standard weights.
Then you make your regression line with std 1, then you calculate std 2 against this regression line.
Then you can use %deviation to see the difference

Ace
Hi ,
I figured this out and using a custom field to calculate std verifcation
thanks
Hi ,
I figured this out and using a custom field to calculate std verifcation
thanks
Do you mind to share your custom field?

Ace
8 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry