Hey all, so the title pretty much says it all... I'm processing everything just fine and everything is correct, but for some reason everything is processed twice, so I have a duplicate of every result, this is especially annoying when reporting tables. Any ideas on why this might be? I've never had this problem until it seemingly came out of nowhere. Thanks in advanced!
"Normal" in the function spot on ea. line of the sample set method followed by a "Quantitate" line at the end can do this. Get rid of one or the other...
Thanks,
DR
Image
"Normal" in the function spot on ea. line of the sample set method followed by a "Quantitate" line at the end can do this. Get rid of one or the other...
Ah, so if I change "Normal" to "Do not process or report" I should be fine then? Thanks!
Right. Same thing applies to standards - if they're processed "Normal", tehn you do not wan a "Calibrate" line at the end, or they'll be doubled as well.
Thanks,
DR
Image
4 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry