Does anyone know how to increase the sensitivity of Naphthalene in a low level (0.5 ppb) standard? I am running EPA method 524 for VOCs, and currently seeing only ion 128 and a signal barely above background noise. I am using an Agilent 6890N GC, Agilent 5973inert GC, 25 mL purge volume, inlet temp 130 C, aux temp 200 C, and a split ratio of 1:10. Any suggestions would be greatly appreciated.
-
- Posts: 43
- Joined: Fri Apr 15, 2011 8:04 pm
Without seeing your other parameters, that inlet temp looks low. What temps are your oven and transfer line at? You want to go low -> high from the sample to the inlet, otherwise you can condense things like naphthalene and the less volatile constituents. We run 524.2 at 100 to 150:1 split ratios and still see down to 0.1-0.2 ppb relatively easily.
Nathan Valentine
Purge and Trap Product Line Manager
Teledyne Tekmar
http://www.teledynetekmar.com
http://www.teledynetekmarblog.com/
Purge and Trap Product Line Manager
Teledyne Tekmar
http://www.teledynetekmar.com
http://www.teledynetekmarblog.com/
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
For naphthalene all you will see is 128 ( that's why the method does not include a qualifier ion).
25 ml purge with a 1:10 split what size column are you using?
I purge 25 ml, split 150:1 and use a 0.18 column.
Beware water.
25 ml purge with a 1:10 split what size column are you using?
I purge 25 ml, split 150:1 and use a 0.18 column.
Beware water.
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
On my 5973 I can easily see a .5 ppb std even using the 4 minute desorb (reduced for 524.3) with 5 mL purged and a 50 : 1 split.
A 1 : 10 split with 25 mL purge and a 4 minute desorb will drown your ion source in water.
A 1 : 10 split with 25 mL purge and a 4 minute desorb will drown your ion source in water.
4 posts
Page 1 of 1
Who is online
In total there are 69 users online :: 0 registered, 0 hidden and 69 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 69 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 69 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science