Steps that I have done to try to correct the problem
I have checked the front inlet and the column for leaks and none were present.
I have cleaned the injector port. the port itself is new and should not need replacing.
I have restarted chemstation
I tried to load the default method to see if the current method we are running is corrupted.
I have replaced the column nut, reducing nut, splitless disk and other routine maintenance.
I looked up Agilent's manual and tried to zero the calibration for the front inlet pressure, flow, and septum purge.
I followed the manual as best I could so if someone thinks this may be the problem I will be more than willing to try and zero again.
If there are any questions please don't hesitate to ask. Any help/advice would be much appreciated as I am still learning the ins and outs of the GCMS.