Degasser starving pump at 1 mL/min
Posted: Wed Feb 27, 2013 5:41 am
We have a MetaChem Degassit (model 6414) which I've just discovered is starving the pump of mobile phase and causing a significant drop in flow rate (85% of 1 mL/min). The problem is definitely the degasser, as I've fed the mobile phase directly into the pump without the degasser and the flow rate became 1 mL/min. (Mobile phase is 50% MeOH in water)
I've put a new inlet filter in the mobile phase reservoir; both the degasser reservoir sit a good 40 cm above the pump, and I've even tried raising the reservoir a further 30 cm higher. Also tried different plumbing configurations on the degasser. Nothing improves the feed to the pump. I've taken the lid off the unit and can see nothing amiss.
Is there anything else I can try or do I just need to buy a new degasser?
I've put a new inlet filter in the mobile phase reservoir; both the degasser reservoir sit a good 40 cm above the pump, and I've even tried raising the reservoir a further 30 cm higher. Also tried different plumbing configurations on the degasser. Nothing improves the feed to the pump. I've taken the lid off the unit and can see nothing amiss.
Is there anything else I can try or do I just need to buy a new degasser?