I found a massive air leak after changing liners and septum and have re installed it but the lick though reduced is still there. What do I do please someone help me.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
To better check if the leak is coming from the inlet try checking the air at higher and lower carrier flows(low oven temperature though to prevent damage to the column from the air). If the air goes up with higher flows it could be somewhere before the injection port, not just at the injection port.
What manufacturer is the GC?
If Agilent, the gas lines going into the injector top can get very tiny cracks where they pass through the shield attached to the top and can be difficult to diagnose until they become larger.
Another test you can perform is capping off the column by pushing the injector end into a septa and see if the air becomes less and the source pressure drops, if it doesn't then you could have a broken column.
If the column is new, and has just recently been installed, check the nut where the column enters the MS as those tend to shrink in the first few days/oven cycles when new. This can often appear as a high air background at low oven temperatures that goes away as you increase the oven temperature during an analysis cycle.
Good luck, I hope it is a simple fix
What manufacturer is the GC?
If Agilent, the gas lines going into the injector top can get very tiny cracks where they pass through the shield attached to the top and can be difficult to diagnose until they become larger.
Another test you can perform is capping off the column by pushing the injector end into a septa and see if the air becomes less and the source pressure drops, if it doesn't then you could have a broken column.
If the column is new, and has just recently been installed, check the nut where the column enters the MS as those tend to shrink in the first few days/oven cycles when new. This can often appear as a high air background at low oven temperatures that goes away as you increase the oven temperature during an analysis cycle.
Good luck, I hope it is a simple fix
The past is there to guide us into the future, not to dwell in.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
Is the reducing nut tight enough to seal the gold seal. Is the o-ring new.
How about the column is the column sealed with the graphite ferrule. Are you sure it is the inlet?
How about the column is the column sealed with the graphite ferrule. Are you sure it is the inlet?
-
- Posts: 2
- Joined: Fri Jan 04, 2013 8:16 am
Thank you all. I found that the column nut got loosen as a result of the cooling process before the septum and liner change. So tightening it solved the problem( the two nuts).
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Air leak at the GC - MS interface is not uncommon after an oven cool down. Consider Sil-Tite ferrules.
5 posts
Page 1 of 1
Who is online
In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 33 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 33 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science