alcohols and glycols
Posted: Fri Feb 22, 2013 10:27 pm
by Sallybeetle
Can anyone recommend general purpose capillary GC column(s) for alcohols and glycols? Also, if you would direct me to application notes for these compounds, it would be most appreciated. I will be using autoinjector or headspace and have a split / splitless 6890 GC with FID and 6890 with MSD.
Thank you.
Re: alcohols and glycols
Posted: Fri Feb 22, 2013 11:59 pm
by chromatographer1
Do a search on the forum. You will find that 1301/624 capillaries and PEG capillaries are the two most popular phases for alcohol and glycol work.
best wishes,
Posted: Sat Feb 23, 2013 2:25 am
by sllimbri65
RTX-WAX column
Posted: Sat Feb 23, 2013 3:45 am
by chromatographer1
Yes, that is a PEG column.
Thanks.
Posted: Tue Mar 05, 2013 8:07 pm
by twranger
We started doing this work about a year ago, and there is very little information for these compounds. Most of the supplier chromatograms are in a solvent or percentage levels, not trace levels in water as we are testing. Best information I found was related to clinical work, using derivitization headspace or direct inject. We are getting more requests for glycols and glycol ethers lately, so we are still researching these!
Posted: Tue Mar 05, 2013 11:19 pm
by chromatographer1
These compounds can be done by HPLC with RI detector at very low levels (< 1ppm) if large sample injection techniques are used, usually when dissolved in water or tested from water using 100% water as the mobile phase.
di ol and tri ol compounds are often done by derivatives, possibly using HS analysis.
These glycols can be a challenge. I hope you enjoy your experiences as I did mine.
best wishes,
Posted: Wed Mar 06, 2013 4:59 am
by jss37992
I use the Phenomenex ZB-WaxPlus for all of my alcohol/glycol work. I have used the Rtx-Wax before and got equal results, the Phenomenex was just a bit cheaper when I bought it.