The method allows for the use of MS, provided you can demonstrate satisfactory performance of the method on your MS. As to whether NELAP will allow you to use a triple quad over a single quad, that might be a question for them to answer.
http://www.nelac-institute.org/interpret-request.php I'd do my MDL/LOD/LOQ verifications using both techniques to make sure you can meet the method requirements using both techniques.
http://water.epa.gov/scitech/methods/cw ... cs_608.pdf
Section 1.2:
"...Method 625 provides gas chromatograph/mass spectrometer
(GC/MS) conditions appropriate for the qualitative and quantitative confirmation of
results for all of the parameters listed above, using the extract produced by this method."
"12.1 Table 1 summarizes the recommended operating conditions for the gas chromatograph.
Included in this table are retention times and MDL that can be achieved under these
conditions. Examples of the separations achieved by Column 1 are shown in Figures 1
to 10. Other packed or capillary (open-tubular) columns, chromatographic conditions,
or detectors may be used if the requirements of Section 8.2 are met."