Separation of two peaks with the same retention times

Posted: Thu Feb 21, 2013 2:10 pm
by josephip
I am currently trying to seperate two chromographic peaks which have the same retention times. I use two mobile phases with 95% and 90% methanol. Does anyone have any tips?

Re: Separation of two peaks with the same retention times

Posted: Thu Feb 21, 2013 3:04 pm
by tom jupille
You're not giving us a whole lot to go on here.
What do you know about the molecules (how do they differ)?
What k' are you getting with 90‰ MeOH?
What column?
What temperature?
What pH?
What else have you tried?

Re: Separation of two peaks with the same retention times

Posted: Mon Mar 04, 2013 4:38 am
by aldehyde
spectral analysis with a DAD or two columns. Or change your method conditions, either way we need way more info about your equipment and sample.

Re: Separation of two peaks with the same retention times

Posted: Thu Mar 07, 2013 11:49 am
by Gerhard Kratz
If one of your compounds is neutral and the other charged you can try HILIC columns to get a baseline separation. Are both neutral but have large difference in MW SEC can be tried.
Anyhow, more information is needed to help you more.