- Posts: 11
- Joined: Fri Jan 18, 2013 3:51 pm
1. I have not be able to find a diluent that will achieve 100% recovery of the material other than 100% MeOH, which does not match up with the mobile phase (65% ACN, 35% H2O * currently) it was (90% ACN/10% H2O).
2. The lower percentage of organic used the lower the response of the API as well as a much broader peak is achieved.
3. I have tried both C18 and CN columns with different dimensions to ensure that resolution can occur between any degradant peaks. On either column I still run into the first 2 issues. I achieved better resolution with higher water content but I sacrifice peak shape. I have also tried gradients that have been found to not work as well.
4. A quick test was ran to see if the API was dissolved and diluted in 100% MeOH first, then diluted 1:1 with water (Final Diluent - 50% MeOH/50% H2O) would give better peak characteristics. I have ran this test multiple times and have continuously ran into the issue of the peak shape deteriorating if lower organic was used. This time when the organic % was lowered the peak shape looked great for both diluents, but now I am seeing a strange shift/jump or drop in the baseline after 20 minutes of run time. This strange shift only occured with the samples that were prepared in 100% MeOH but was not consistant, while the 1 sample with 50% MeOH/50% H2O as its final diluent gave a flat baseline.
I am at the point to where I do not know what to try next, Any thoughts or ideas that may help?