How to calculate volume of stationary phase in a column?
Posted: Mon Feb 18, 2013 9:41 pm
by lsreedharan
Hello all,
I calculated the total volume of my column. But, Im not sure to calculate the stationary phase volume and mobile phase volume. I'm using 3.0 * 150 mm * 3.5 microns.
Sree
Re: How to calculate volume of stationary phase in a column?
Posted: Tue Feb 19, 2013 1:10 am
by tom jupille
Do you want to *calculate* it or *estimate* it?
If you want to estimate, for reversed phase columns the volume of mobile phase (Vm) is approximately 0.5 x L x dc^2 where L is the column length and dc is the column ID. This is usually correct within about +/-15% or so.
If you want to calculate, you can't; you have to measure the dead time (t0) and then multiply by the flow rate to get Vm. Getting an "accurate" measure of t0 is itself a complicated issue; for example this discussion from the early days of the Forum:
http://www.lcresources.com/discus/messa ... 3/215.html
Posted: Tue Feb 19, 2013 3:27 am
by lsreedharan
I was not correct. I need to estimate.
http://people.rit.edu/pac8612/webionex/ ... e4o85.html
I was bit confused after seeing the above link. Thanks Tom.
Posted: Tue Feb 19, 2013 2:57 pm
by KM-USA
Agilent typically states 0.6 x the column radius squared x pi x length due to stationary phase particles
Posted: Tue Feb 19, 2013 3:13 pm
by Peter Apps
The Agilent equation gives the volume of the support - isn't the stationary phase strictly the almost monomolecular layer of e.g. C18 stuck onto the surface of the support ?, or even more strictly still the "outer" surface of the layer, which is the only bit that the analytes get to see ?
Ah the simple joys and geometry of capillary GC :lol: