Separation of two peaks with the same retention times

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
I am currently trying to seperate two chromographic peaks which have the same retention times. I use two mobile phases with 95% and 90% methanol. Does anyone have any tips?
You're not giving us a whole lot to go on here.
What do you know about the molecules (how do they differ)?
What k' are you getting with 90‰ MeOH?
What column?
What temperature?
What pH?
What else have you tried?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
spectral analysis with a DAD or two columns. Or change your method conditions, either way we need way more info about your equipment and sample.
If one of your compounds is neutral and the other charged you can try HILIC columns to get a baseline separation. Are both neutral but have large difference in MW SEC can be tried.
Anyhow, more information is needed to help you more.
Gerhard Kratz, Kratz_Gerhard@web.de
4 posts Page 1 of 1

Who is online

In total there are 349 users online :: 1 registered, 0 hidden and 348 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 348 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry