Unusual small tailing peaks

Posted: Sun Feb 17, 2013 10:41 pm
by eduardorickrot
I'm an inexperienced (eight months on a 20 hours per week job) trainee working at a quality control laboratory in Brazil.
Last Friday, while doing routine work, a small and highly tailed, but significant peak started appearing in our analyzes from one of the FID chromatographs, at around 4 minutes in a 6 min. run, in both samples and patterns. I found it very strange, because I haven't seen anything like that before, although it didn't interfere much with our work, so no one cared much about it, but I'm still curious. Any ideas about the cause?

Re: Unusual small tailing peaks

Posted: Mon Feb 18, 2013 12:05 pm
by TiagoBele
E aí Eduardo! Tudo certo?

Já me deparei com esta situação anteriormente com um gs/ms Agilent, 7890A/5975C. Vc sabe se foi realizada alguma manutenção de rotina antes de começar a injetar?
Geralmente trocando o septo do injetor e/ou o liner resolve seu problema. O fênomeno ocorre quando forma-se sítios ativos no liner ou quando alguma partícula (sim... bem pequeno mesmo... quase invísivel a olho nu) do septo passa para o liner, aumentando a retenção dos compostos injetados nestes sítios, e assim distorce a sua simetria de pico.
Se o problema continuar considere cortar um pedacinho da coluna (de 2 a 4 centímetros sempre dá certo comigo) que está ligada no começo do injetor (a sujeira pode sim ter passado pra frente). Ela sangrará logo após, mas nada que um bom condicionamento ou umas 5 corridas dê uma boa limpada.

Qualquer coisa entre em contato! tiago_bele@terra.com.br

Abraços!

Tiago G. A. Belé