Negative signal of the GC: I am using Hewlett 5890 Series II Gas Chromatography with FID detector and fused silica column (Rtx-200). It is giving negative signal with or without injection. Beginning of the run the signal is positive then turns to negative; during the running time the signal does not go back to positive, after end of the run the signal goes back to positive.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
It sounds like your FID is detecting a bleed from your column which is decreasing with time. It then resets the baseline at the next start of run (autozero function)
Your bleed could be from the column or from somewhere upstream.
best wishes,
Rod
Your bleed could be from the column or from somewhere upstream.
best wishes,
Rod
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Are you running the carrier gas a constant flow or constant pressure ? If at constant pressure the volume flow rate will decrease as temperature increases, but this effect is usually cancelled by bleed from the column.
Peter
Peter
Peter Apps
-
- Posts: 9
- Joined: Fri Jan 25, 2013 10:58 pm
Thanks for responses.
I tested with different column but still the signal is negative.
The carrier gas flow is constant.
I tested with different column but still the signal is negative.
The carrier gas flow is constant.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
By the signal going to negative, do you mean that it decreases below the original zero, or that you see an upside down chromatogram ?
How fast does this change to a negative signal happen, and how far into the run ?
Peter
How fast does this change to a negative signal happen, and how far into the run ?
Peter
Peter Apps
-
- Posts: 9
- Joined: Fri Jan 25, 2013 10:58 pm
No, I did not see upside down chromatogram but base line shifted below to the original base line. For example at the beginning of the run (oven temperature = 80 C) the signal is +3.4 after 15 min the signal is -95 (oven temperature=210).
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
If you have the gas saver activated try deactivating it.
Peter
Peter
Peter Apps
-
- Posts: 9
- Joined: Fri Jan 25, 2013 10:58 pm
Thanks for your response. Our GC does not have gas saver.
-
- Posts: 350
- Joined: Tue Jun 28, 2011 7:30 pm
I would suspect a leak somewhere that goes away as the oven heats up.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Either carrier gas contamination (a rather volatile one) from upstream or from a lubricant in the GC or a cracked column could cause the problem described. A problem in the FID itself is a less likely but possible issue.
good hunting,
Rod
good hunting,
Rod
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
You are working with constant carrier flow - what is the flow setting (in ml/min or cm/s) ?
Peter
Peter
Peter Apps
-
- Posts: 9
- Joined: Fri Jan 25, 2013 10:58 pm
Column flow= 2.9ml/min
He + column flow= 48.8ml/min
Air + column flow= 320.9ml/min
H2 + column flow= 43.3ml/min
He + column flow= 48.8ml/min
Air + column flow= 320.9ml/min
H2 + column flow= 43.3ml/min
-
- Posts: 9
- Joined: Fri Jan 25, 2013 10:58 pm
I removed the column from the detector side. Seal the side of the detector. Start isothermal run: As soon as I press “Start”, base line moves to original zero to negative side than if I press, “STOP” it go back to original base line.
All over the process temperature, gas flow all other condition are same only I pressed start.
Any more idea…………
All over the process temperature, gas flow all other condition are same only I pressed start.
Any more idea…………
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
I would guess that your pneumatics are contaminated and the bleed is slowly decreasing with time.
Time to perform a thorough cleaning of your pneumatics.
best wishes,
Rod
Time to perform a thorough cleaning of your pneumatics.
best wishes,
Rod
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
chromatography1,
Do you have column comp on? Do you have a zero set on your GC?
Best regards,
AICMM
Do you have column comp on? Do you have a zero set on your GC?
Best regards,
AICMM
21 posts
Page 1 of 2
Who is online
In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 26 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 26 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science