Page 1 of 1

storage standar vials

Posted: Sat Feb 09, 2013 5:31 pm
by pueblito
Dear Chromatographers.

I have received some standard ampules indicating :

Diesel fuel.... must be stored at 2 - 8 °C

Arochlor 1232.... should be stored at 5 °C

Aromatic Hydrocarbon Std... must be stored 20 - 30 °C.

I use to break the ampule, then transfer into a vial and I store the vial in a freezer at around -2 to -4 °C, but now, I donk know what shoul I do?

Any suggestion would be appreciated.

bye.

Re: storage standar vials

Posted: Mon Feb 11, 2013 1:31 pm
by Don Shelly
I recommend asking the company that you purchased them from. Some analytes will precipitate out of solution at cold temperatures. Generally, careful sonication puts them back into solution.

Re: storage standar vials

Posted: Tue Feb 12, 2013 4:36 pm
by pueblito
Don, Thanks for your help.

Re: storage standar vials

Posted: Tue Mar 05, 2013 8:13 pm
by twranger
We have stored ours in a refrigerator (reguardless of recommendation) for years with no problems. Freezing is overkill for anything except volatiles. Most of our semivols recommend room temp.

Re: storage standar vials

Posted: Sun Mar 10, 2013 6:02 pm
by pueblito
So, maybe storaging at low temperatures is not a good idea after all.

Re: storage standar vials

Posted: Wed Mar 27, 2013 6:23 pm
by Bigbear
Interesting. We use a custom mix for a few of our 525 compounds. We buy 2 concentrations as one is used for a second source. The low concentration is to be stored in a freezer and the higher concentration in the fridge!!

Re: storage standar vials

Posted: Wed Mar 27, 2013 6:41 pm
by Steve Reimer
Just a guess but, the larger PAHs may have a solubility problem at freezer temp and higher concentration.