Dear All,

During the investigation, we identified the peak showing in Water (Diluent) Injection in HPLC with method having @210nm. Its keep increasing over the time. Does anyone seen this kind of pattern. Your suggestions will be helpful.
Is this a gradient method? If yes, does the peak appear when running a no-injection blank?
Yes, its gradient method. I cannot see the peak in gradient(no injection). The peak only appeared in blank (water) & subsequent injections. When we inject the same blank vial (without cap) multiple injections, the unknown impurity increasing. I dont understand the phenomenon that how can pure water can give the unkwon peak at same retention time?
I would guess that something is being absorbed into the sample after it is prepared. Maybe something volatile in the lab atmosphere?
4 posts Page 1 of 1

Who is online

In total there are 356 users online :: 2 registered, 0 hidden and 354 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 354 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry