What do you consider for method verification protocol
Posted: Thu Feb 07, 2013 3:23 pm
I am here to see your opinions:
IF you have experience transfering method and doing verification based on the validation protocol and report.
For a validated HPLC method (impurities/assay)
to transfer it to another lab,
what parameters would you consider in the verification protocol?
Just got USP35: Chaper <1224> Transfer of Analytical Procedures
There are several recommended approaches:
1) Comparative Trsting (test the same samples)
2) Covalidation btw 2 labs
3) Revalidation or partial revalidation
4) Transfer waiver
There are two voices in this forum 1) and 3)
I am the 3) side
system suitability, specificity, LOD/LOQ, Linearity, range, accuracy, repeatability, ruggedness (intermediate precision).
Since these parameters depend on instrument/operator.
I’d not do stability/robustness since this is the intrinsic character of the API/ DS/DP and the method.
This covers almost everything in the validation, but with the validation results, it is much easier to do.
IF you have experience transfering method and doing verification based on the validation protocol and report.
For a validated HPLC method (impurities/assay)
to transfer it to another lab,
what parameters would you consider in the verification protocol?
Just got USP35: Chaper <1224> Transfer of Analytical Procedures
There are several recommended approaches:
1) Comparative Trsting (test the same samples)
2) Covalidation btw 2 labs
3) Revalidation or partial revalidation
4) Transfer waiver
There are two voices in this forum 1) and 3)
I am the 3) side
system suitability, specificity, LOD/LOQ, Linearity, range, accuracy, repeatability, ruggedness (intermediate precision).
Since these parameters depend on instrument/operator.
I’d not do stability/robustness since this is the intrinsic character of the API/ DS/DP and the method.
This covers almost everything in the validation, but with the validation results, it is much easier to do.