Hello everybody, I wanted to ask something about an HPLC LKB, the pump is a 2150 model. Recently I accidentally flushed the system with acetonitrile, and before that I was running a mobile phase with phosphoric acid 0.2%:acetonitrile (9:2). Anybody has an idea about how to clean the valves? I did it with some new equipments but I wanted to know if anybody had such a problem with an equipment that old.
Thank you very much in advance.
Generally speaking, remove the check valves and sonicate in methanol.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
For phosphate I would use a mixture of methanol and water.
I am not sure about the model but our LKB pumps at Uni you could dissasemble the check valves, If you do that do not take your eyes of the ball. The ruby ball seals are extremely difficult to find once it drops on the floor, trust me.
3 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry