EP Impurity Method for Timolol maleate is not robust
Posted: Wed Feb 06, 2013 8:33 am
Dear Colleagues,
I am currently doing a robustness study of pharmacopendial methods and noticed that the EP impurity method for Timolol maleate is not robust regarding the retention time of the API peak Timolol maleate.
Under nominal conditions my peak comes at approximately 7 minutes but when I vary the amount of methanol by +/- 10% in eluent A, the retention times shift to 30 and 2 minutes respectively. This I find a bit to extrem and is not robust enough for daily routine testing.
The method details are listed in the current EP and are as follows:
Column: 125 x 4 mm, 5 um Octadecylsilicate. The column I use is a Nucleosil C18 von Machery & Nagel with the same dimensions.
Buffer: A sodium octane sufonate solution adjusted to pH 3.0 with 99% acetic acid. I apply a 4.32 g/L sodium octane sulfonate solution adjusted to pH 3.0 with glacial acetic acid.
Mobile Phase A: 50:50, Buffer: Methanol (v/v)
Mobile Phase B: Methanol
Gradient:
0 -10(min) 97.5%A 2.5%B
10 - 11(min) 97.5%A to 70%B ; 2.5% A to 30%B
11 - 14.5(min) 70%A 30%B
Flow rate: 2.0 ml/min
Detection: 295 nm
Injection volume: 20 ul
I varied the pH by +/- 10% and it had no significant effect on the retention time of the Timolol maleate peak. My question therefore is, could this extrem shift (instability) in retention time be due to my Nucleosil C18 column, methanol as eluent/solvent and/or the Mobile phase ratio of 2.5% to 97.5% during the first 10 minutes?
Many thanks in advance!
Max P.
Gradient:
Detection: 295 nm
Max P.