GC-MS 6890/5973N

Posted: Tue Feb 05, 2013 9:33 pm
by Archimonde1308
How reliable is scan mode for quantifying components? I have too many compounds in my sample and setting up a SIM mode analysis is turning out to be a real nuisance.
Thanks in advance!

Re: GC-MS 6890/5973N

Posted: Wed Feb 06, 2013 6:27 pm
by Yama001
There is no difference as long as you have sufficient sensitivity in scan mode. I would prefer scan in any application unless I needed the improved signal to noise of SIM.

Re: GC-MS 6890/5973N

Posted: Sun Feb 10, 2013 11:40 pm
by MSCHemist
If you have adequate sensitivity you can analyze each compound with a specific quant ion off the scan as you would in SIM rather than the TIC area. Back in my last position I did 42 compound in citrus juices this way.