low level rl's for naphthalene using purge and trap

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

8 posts Page 1 of 1
Any luck with running naphthalene meeting a RL of 0.5ppb? I am going to try a larger volume purge and see what I can do, but just curious.

Agilent 6890
Encon purge and trap
Archon autosampler

Will be trying on the soil side first.
I can get enough response to see 0.5ppb with no problem if I am only looking for the 128 peak. The problem is that I get nasty samples and naphthalene is very persistent. With my system (6890/5975 OI4660 - Archon) a larger purge wouldn't be as helpful as careful screening.

The response is usually quadratic at low levels.
I have the same problem with the samples - unfortunately even though some of these samples have to be diluted for naphthalene the client still wants a RL of 0.5ppb without running it by EPA8270. I ran a 10ml purge last night with an ical only going to 25ppb so I'm hoping to achieve the goal.
I'll concur with Steve - detecting naphthalene at 0.5 ug/L is doable - not detecting it in clean samples is the problem.
Great sounds good, thanks. These are definitely not clean
Works for me. Do 524 see down to 0.5 ppb. purge 25ml.
I have the same setup and can see 0.5ppb with little problems. We purge 5ml sample.

What type of trap are you using?
What is your split ratio?
What GC column?



I use a custom packed trap with Carboseive SIII, Carboxen1000 and CarbopackB. Similar to the VoCarb3000. For Naphthalene use a Vocarb3000 or Vocarb4000 or possibly the BTEX trap. The biggest problem I have so far is the carryover of Naphthalene, but if you run the transfer lines at 130c and use the 5ml sparge tube and rinse with 8ml DI after a 5ml sparge it helps.

Been running purge and trap for twenty years and still there isn't a day I don't see some new problem with it. Just keeps things exciting :)
The past is there to guide us into the future, not to dwell in.
Thanks James_Ball

I use the Voacarb 3000, DB624 column with a 1:50 split. I was actually able to see it with the 5ml purge and get a decent result +-30% after chopping off most of my upper points and just leaving 6 points with a quadratic curve.

I will have a really hard time of it if some of site gets cleaned up since the carryover will be awful. So far I have never reported anything below 5ppb for them.

I have only been running purge and trap for ~8 years now, but it is my favorite since it never gets old. :P
8 posts Page 1 of 1

Who is online

In total there are 76 users online :: 0 registered, 0 hidden and 76 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 76 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry