Fatty Acid Methyl Ester, very late peak elution?

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
First, a bit description on how I prepared the smaples, then move on to GC.
A. Lipid extraction:
On 0.1ml animal blood cells
1. add 0.1 water, incubate for 15 minutes.
2. add 2.2ml isopropanol containing 0.005%BHT, place tube on shaker for 1h.
3. add 1.75 ml chloroform, votex and place tube on shaker for 1h.
4. centrifuge, take supernatant, and dry under nitrogen, reconstitute in 0.25ml hexane.

B. Tranesterification:

add 2.5ml freshly prepared 73%methanol:18%hexane:9%acetyl chloride (volume%), sealed, and heated at 100C for 1 hour.
Cool, add 4.5ml potassium carbonate, mixing and centrifuge.

C. GC:

inject 0.5 microliter to an Omegwax 250 column (30m x 0.25mm)

Now. I am seeing some peaks extremely late.
Image

The 22:6n3 is supposed to be the last peak on this column with a retention time of 45 minutes. However, I am seeing a bunch of peaks after that, some eluted at 125minute!

If I run the sample for 70 minutes, these peaks would carry over to the next run.

What are these peaks? Are they free acids because of incomplete esterfication/transesterification? Are they artifacts? Do you ever see peaks eluting that late? By the way, I ran standards, solvent, and blanks, and these peaks were not there if there were no carryover from the previous runs. So these late peaks are from the real samples. In addition, the retention times of these late peaks were consistent in all samples, so they are real peaks.

I suspect that these are polar components because Omegawax has a polar stationary phase, but not sure. I may try extra extractions on the samples, which would get rid of the polar components.

I want to do a direct injection without purification. But if that is not possible because of the nature of samples, do you have a fast and easy purification method to recommend? I heard of passing sample through a silica gel bed in a disposable glass pipette, does it work well?
The only way to know what those could be is if you are running a GC/MS. That way you can differentiate between contamination and a true product. If you are not running a GC/MS you can trim about 10-20 mm from the column end of the inlet. Run several blanks and see if you can establish a clean baseline. You may also want to install a clean GC inlet liner to see if this helps. If I had to guess, it could be siloxanes in the column which is a common contamination in GC. Goodluck
have you done a solvent injection?
Yes, I have done solvent injections. These late peaks were not in the solvents.

I have ordered some silica gels, and will try to pass through the FAME sample through a silica gel column before GC. Will report back if that helps.
4 posts Page 1 of 1

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry