I am fairly new to HPLC and I am trying to understand the interactions between the mobile phase and stationary pahses I use at my work. I would like some insight on the effect that TeA has on a reverse phase C18 column. I have read that TeA acts as a competing base but what are the benefits of this? Also I have been told that TeA is not needed with newer columns, is this a valid statement? Any feedback or literature is much apprecitaed.
Reverse phase supports usually have a silica base (chemically polar with acidic and basic character), which is reacted with chemical attachments which give it a non-polar character. Not all of these reactive sites are blocked with a non-polar functional group and some active sites are embedded metal ions in the silica.

A second chemical reaction is often executed, blocking many of these active sites, but different procedures give differing results. This is also affected by the quality of the silica starting material.

The unblocked acidic sites can be temporarily blocked by an amine like TEA which reduces interactions with an analyte with the silica backbone of the RP support. This blocking reduces the interaction of the acidic sites which may remain on the support which cause 'tailing' of the peaks eluting from the column.

Advancements have been achieved with the chemical reactions blocking these active sites so TEA does not improve many of the newer columns performance.

Unfortunately for older columns the TEA is not always completely removed when purged with a mobile phase which is lacking this chemical addition. Performance then for other samples may not reproduce the column's original performance for these samples. This is not a desirable situation.

These factors are why there are so many different RP C-18 columns available. Older columns are still manufactured using their original silica with all of its 'defects' so analytical work produced years ago can be reproduced and some of these columns may require TEA for the testing they were originally chosen to perform.

Make your choice of a column carefully, and after making your choice it behoves you to ascertain that the column will be available in the future and not discontinued.

You should do some reading and there are many books available to teach you the fundamentals of HPLC.

ENJOY !

Rod
2 posts Page 1 of 1

Who is online

In total there are 27 users online :: 2 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry