Advertisement

PAHs analysis by UPLC - tubings easily clog

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi everyone,

I have a problem when switching the method from Waters HPLC system to UPLC system for PAH analysis. When analysing the sample (dissolved in 100% acetonitrile and filtered thru 0.2um filter) using the UPLC system (with initial gradient of 50/50 ACN/water), the tubing is clogged easily, probably because of the precipitation of some matrices in the sample (in ACN) when mixed with the mobile phase (50% ACN/water). However, I couldn't dissolve my sample in 50/50 ACN/Water at the start because some PAHs may not dissolve well in this solvent. Is there any solution for this? Please advice.

Hub
Have you verified that it is, in fact, the tubing clogging rather than the guard cartridge or column inlet frit?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Hi Tom,

No. However, what The set of columns used in UPLC is exactly the same as when I use in the HPLC. Thereofore it is just the wise guess. Also, after replacing some of the tubings, the pressure resumes normal.

Hub
Also, after replacing some of the tubings, the pressure resumes normal
That pretty much verifies that it's the tubing. :wink:

The first thing I would do is to install a 0.2 micron in-line filter immediately downstream of the injector. Those are designed to be easy to change.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 27 users online :: 2 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry