Carry Over??

Posted: Thu Jan 31, 2013 12:32 pm
by amishshah
Do control carry over allows in assay analysis?

Re: Carry Over??

Posted: Fri Feb 01, 2013 2:12 pm
by JMcK
Generally you want to get all the analytes off in single run. If you're having carryover it could coelute with your analyte of interest. I don't know how much freedom you have to adjust the method. If allowed, you can extend the run to ensure all the analytes elute, or, if it's a gradient reverse phase method, increase the organic phase concentration at the end of the run to push off the remaining analytes. You may have to equilibrate longer after each injection. Just some thoughts.

Re: Carry Over??

Posted: Mon Feb 04, 2013 9:00 pm
by zheyin
Generally you want to get all the analytes off in single run. If you're having carryover it could coelute with your analyte of interest. I don't know how much freedom you have to adjust the method. If allowed, you can extend the run to ensure all the analytes elute, or, if it's a gradient reverse phase method, increase the organic phase concentration at the end of the run to push off the remaining analytes. You may have to equilibrate longer after each injection. Just some thoughts.
These are not carryover.

Re: Carry Over??

Posted: Mon Feb 04, 2013 9:01 pm
by zheyin
Do control carry over allows in assay analysis?
I'd allow < LOD detection in carryover tolerable in my SOP.

Re: Carry Over??

Posted: Tue Feb 05, 2013 1:36 pm
by JMcK
Do control carry over allows in assay analysis?
These are not carryover.

I'd allow < LOD detection in carryover tolerable in my SOP.
I'm not sure what you're talking about then, and I doubt I'm the only one.