Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
I have some trouble with my Agilent 1200 RID. CS A.10.02. Running 0.6 mL/min with 10 mmol sulfuric acid. Regardless of what pumps there seems to be some electronic
problem. Take a look at the pic below.
Image
The RID signal fluctuates by about 60,000 units. Turning the RID heater on and off seems to make no difference. Note that the temp of the optical unit seems to spike at
the height of it's ripple but the periodicity of this is constant whether optical temp is on or not. Diode1 and diode2 both spike right after the temp spike. Any ideas?

Thanks,
Marc
Hi Marc,
it looks like there are air bubbles in your system. Please check the RI flow cell first.
Good luck.
Gerhard Kratz, Kratz_Gerhard@web.de
Run with pump at zero flow, to confirm that the issue is with the pump.

What do you mean by regardless of what pumps? Do you mean different pumps or just different channels of a quat pump? Are you bypassing any solvent mixing valves for your mobile phase?

And what about basics, purging, degassing, etc.?
Just wanted to add that the pressure is perfectly constant at 60 bar, there is no injection and the UV is monitored
at 200 nm (measured just before the RID) and the UV is also completely flat (<0.01 AU noise and flat and random)
My wild guess is that you have a bit of outgassing (or, depending on the flow resistance of the RI cell, maybe a small leak at its inlet such that it can aspirate air), with a small bubble hanging in the flow cell and accumulating friends to gradually grow larger until it finally breaks loose to start the cycle anew. The fact that you don't see it on a UV detector *before* the RI may simply be that either a) there is enough back pressure on the UV cell that the bubbles don't form until the outlet or b) the bubbles are small enough not to show up as spikes and the UV flow cell geometry is such that they don't hang up and accumulate.

Following up on Gerhard & KM's comments: cut the flow in half and see if the period (approximately) doubles.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts Page 1 of 1

Who is online

In total there are 43 users online :: 1 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry